1. How should the integrity of sterilising filters be verified? H+V June 2007
Annex 1, paragraph 85 states, 'the integrity of the sterilised filter should be verified before use and should be confirmed immediately after use by an appropriate method such as a bubble-point, diffusive-flow or pressure-hold test.'
The filter-sterilisation process may be physically stressful for the filter. For example, high temperatures during the process may cause the filter to distort, potentially leading to fluid pathways that allow the passage of particles greater than 0.2 μm in size. The performance of a filter can improve with use, as particles begin to block individual pathways and remove larger pathways that smaller particles could successfully navigate. For these reasons, filters should be tested both before use but after sterilisation and again after use.
Furthermore, testing should be performed in situ in order to verify the integrity of the filter complete with its housing.
蒸汽灭菌和安装过程中都纯在完整性破坏的风险。 |
|