发布部门:FDA发布国家:美国发布日期:1997年3月11日生效日期:1997年3月11日原文链接:http://www.fda.gov/medicaldevice ... ments/ucm070627.htm全文贴出
Design Control Guidance For Medical Device Manufacturers

This Guidance relates to FDA 21 CFR 820.30 and Sub-clause 4.4 of ISO 9001
March 11, 1997
FOREWORD
To ensure that good quality assurance practices are used for the design of medical devices and that they are consistent with quality system requirements worldwide, the Food and Drug Administration revised the Current Good Manufacturing Practice (CGMP) requirements by incorporating them into the Quality System Regulation, 21 CFR Part 820. An important component of the revision is the addition of design controls.
Because design controls must apply to a wide variety of devices, the regulation does not prescribe the practices that must be used. Instead, it establishes a framework that manufacturers must use when developing and implementing design controls. The framework provides manufacturers with the flexibility needed to develop design controls that both comply with the regulation and are most appropriate for their own design and development processes.
This guidance is intended to assist manufacturers in understanding the intent of the regulation. Design controls are based upon quality assurance and engineering principles. This guidance complements the regulation by describing its intent from a technical perspective using practical terms and examples.
Draft guidance was made publicly available in March, 1996. We appreciate the many comments, suggestions for improvement, and encouragement we received from industry, interested parties, and the Global Harmonization Task Force (GHTF) Study Group 3. The comments were systematically reviewed, and revisions made in response to those comments and suggestions are incorporated in this version. As experience is gained with the guidance, FDA will consider the need for additional revisions within the next six to eighteen months.
The Center publishes the results of its work in scientific journals and in its own technical reports. Through these reports, CDRH also provides assistance to industry and to the medical and healthcare professional communities in complying with the laws and regulations mandated by Congress. These reports are sold by the Government Printing Office (GPO) and by the National Technical Information Service (NTIS). Many reports, including this guidance document, are also available via Internet on the World Wide Web at www.fda.gov.
We welcome your comments and suggestions for future revisions.
| /signed/ | | | | D. Bruce Burlington, M.D. | | Director | | Center for Devices and Radiological Health |
PREFACE
Effective implementation of design controls requires that the regulation and its intent be well understood. The Office of Compliance within CDRH is using several methods to assist manufacturers in developing this understanding. Methods include the use of presentations, teleconferences, practice audits, and written guidance.
Those persons in medical device companies charged with responsibility for developing, implementing, or applying design controls come from a wide variety of technical and non-technical backgrounds--engineering, business administration, life sciences, computer science, and the arts. Therefore, it is important that a tool be provided that conveys the intent of the regulation using practical terminology and examples. That is the purpose of his guidance.
The response of medical device manufacturers and other interested parties to the March, 1996 draft version of this guidance has significantly influenced this latest version. Most comments centered on the complaint that the guidance was too prescriptive. Therefore, it has been rewritten to be more pragmatic, focusing on principles rather than specific practices.
It is noteworthy that many comments offered suggestions for improving the guidance, and that the authors of the comments often acknowledged the value of design controls and the potential benefit of good guidance to the medical device industry, the public, and the FDA. Some comments even included examples of past experiences with the implementation of controls.
Finally, there are several people within CDRH that deserve recognition for their contributions to the development of this guidance. Al Taylor and Bill Midgette of the Office of Science and Technology led the development effort and served as co?chairs of the CDRH Design Control Guidance Team that reviewed the comments received last spring. Team members included Ashley Boulware, Bob Cangelosi, Andrew Lowrey, Deborah Lumbardo, Jack McCracken, Greg O'Connell, and Walter Scott. As the lead person within CDRH with responsibility for implementing the Quality System Regulation, Kim Trautman reviewed the guidance and coordinated its development with the many other concurrent and related activities. Their contributions are gratefully acknowledged.
FDA would also like to acknowledge the significant contributions made by the Global Harmonization Task Force (GHTF) Study Group 3. The Study Group reviewed and revised this guidance at multiple stages during its development. It is hoped that this cooperative effort will lead to this guidance being accepted as an internationally recognized guidance document through the GHTF later this year.
| /signed/ | | | | Lillian J. Gill | | Director | | Office of Compliance |
ACKNOWLEDGEMENT
FDA wishes to acknowledge the contributions of the Global Harmonization Task Force (GHTF) Study Group 3 to the development of this guidance. As has been stated in the past, FDA is firmly committed to the international harmonization of standards and regulations governing medical devices. The GHTF was formed in 1992 to further this effort. The GHTF includes representatives of the Canadian Ministry of Health and Welfare; the Japanese Ministry of Health and Welfare; FDA; industry members from the European Union, Australia, Canada, Japan, and the United States; and a few delegates from observing countries.
Among other efforts, the GHTF Study Group 3 started developing guidance on the application of design controls to medical devices in the spring of 1995. Study Group 3 has recognized FDA's need to publish timely guidance on this topic in conjunction with promulgation of its new Quality System Regulation. The Study Group has therefore devoted considerable time and effort to combine its draft document with the FDA's efforts as well as to review and comment on FDA's subsequent revisions. FDA, for its part, delayed final release of its guidance pending final review by the Study Group. As a result, it is hoped that this document, with some minor editorial revisions to make the guidance global to several regulatory schemes, will be recognized through the GHTF as an international guidance document.
TABLE OF CONTENTS
INTRODUCTION
I. PURPOSE
This guidance is intended to assist manufacturers in understanding quality system requirements concerning design controls. Assistance is provided by interpreting the language of the quality systems requirements and explaining the underlying concepts in practical terms.
Design controls are an interrelated set of practices and procedures that are incorporated into the design and development process, i.e., a system of checks and balances. Design controls make systematic assessment of the design an integral part of development. As a result, deficiencies in design input requirements, and discrepancies between the proposed designs and requirements, are made evident and corrected earlier in the development process. Design controls increase the likelihood that the design transferred to production will translate into a device that is appropriate for its intended use.
In practice, design controls provide managers and designers with improved visibility of the design process. With improved visibility, managers are empowered to more effectively direct the design process-that is, to recognize problems earlier, make corrections, and adjust resource allocations. Designers benefit both by enhanced understanding of the degree of conformance of a design to user and patient needs, and by improved communications and coordination among all participants in the process.
The medical device industry encompasses a wide range of technologies and applications, ranging from simple hand tools to complex computer-controlled surgical machines, from implantable screws to artificial organs, from blood-glucose test strips to diagnostic imaging systems and laboratory test equipment. These devices are manufactured by companies varying in size and structure, methods of design and development, and methods of management. These factors significantly influence how design controls are actually applied. Given this diversity, this guidance does not suggest particular methods of implementation, and therefore, must not be used to assess compliance with the quality system requirements. Rather, the intent is to expand upon the distilled language of the quality system requirements with practical explanations and examples of design control principles. Armed with this basic knowledge, manufacturers can and should seek out technology-specific guidance on applying design controls to their particular situation.
When using this guidance, there could be a tendency to focus only on the time and effort required in developing and incorporating the controls into the design process. However, readers should keep in mind the intrinsic value of design controls as well. It is a well-established fact that the cost to correct design errors is lower when errors are detected early in the design and development process. Large and small companies that have achieved quality systems certification under ISO 9001 cite improvements in productivity, product quality, customer satisfaction, and company competitiveness. Additional benefits are described in comments received from a quality assurance manager of a medical device firm regarding the value of a properly documented design control system:
"...there are benefits to an organization and the quality improvement of an organization by having a written design control system. By defining this system on paper, a corporation allows all its employees to understand the requirements, the process, and expectations of design and how the quality of design is assured and perceived by the system. It also provides a baseline to review the system periodically for further improvements based on history, problems, and failures of the system (not the product)."
II. SCOPE
The guidance applies to the design of medical devices as well as the design of the associated manufacturing processes. The guidance is applicable to new designs as well as modifications or improvements to existing device designs. The guidance discusses subjects in the order in which they appear in FDA's Quality System regulation and is cross-referenced to International Organization for Standards (ISO) 9001:1994, Quality SystemsModel for Quality Assurance in Design, Development, Production, Installation, and Servicing, and the ISO draft international standard ISO/DIS 13485, Quality SystemsMedical DevicesParticular Requirements for the Application of ISO 9001, dated April 1996.
Design controls are a component of a comprehensive quality system that covers the life of a device. The assurance process is a total systems approach that extends from the development of device requirements through design, production, distribution, use, maintenance, and eventually, obsolescence. Design control begins with development and approval of design inputs, and includes the design of a device and the associated manufacturing processes.
Design control does not end with the transfer of a design to production. Design control applies to all changes to the device or manufacturing process design, including those occurring long after a device has been introduced to the market. This includes evolutionary changes such as performance enhancements as well as revolutionary changes such as corrective actions resulting from the analysis of failed product. The changes are part of a continuous, ongoing effort to design and develop a device that meets the needs of the user and/or patient. Thus, the design control process is revisited many times during the life of a product.
Some tools and techniques are described in the guidance. Although aspects of their utility are sometimes described, they are included in the guidance for illustrative purposes only. Including them does not mean that they are preferred. There may be alternative ways that are better suited to a particular manufacturer and design activity. The literature contains an abundance of information on tools and techniques. Such topics as project management, design review, process capability, and many others referred to in this guidance are available in textbooks, periodicals, and journals. As a manufacturer applies design controls to a particular task, the appropriate tools and techniques used by competent personnel should be applied to meet the needs of the unique product or process for that manufacturer.
III. APPLICATION OF DESIGN CONTROLS
Design controls may be applied to any product development process. The simple example shown in Figure 1 illustrates the influence of design controls on a design process.

Figure 1 - Application of Design Controls to Waterfall Design Process (figure used with permission of Medical Devices Bureau, Health Canada)
The development process depicted in the example is a traditional waterfall model. The design proceeds in a logical sequence of phases or stages. Basically, requirements are developed, and a device is designed to meet those requirements. The design is then evaluated, transferred to production, and the device is manufactured. In practice, feedback paths would be required between each phase of the process and previous phases, representing the iterative nature of product development. However, this detail has been omitted from the figure to make the influence of the design controls on the design process more distinct.
The importance of the design input and verification of design outputs is illustrated by this example. When the design input has been reviewed and the design input requirements are determined to be acceptable, an iterative process of translating those requirements into a device design begins. The first step is conversion of the requirements into system or high-level specifications. Thus, these specifications are a design output. Upon verification that the high-level specifications conform to the design input requirements, they become the design input for the next step in the design process, and so on.
This basic technique is used repeatedly throughout the design process. Each design input is converted into a new design output; each output is verified as conforming to its input; and it then becomes the design input for another step in the design process. In this manner, the design input requirements are translated into a device design conforming to those requirements.
The importance of design reviews is also illustrated by the example. The design reviews are conducted at strategic points in the design process. For example, a review is conducted to assure that the design input requirements are adequate before they are converted into the design specifications. Another is used to assure that the device design is adequate before prototypes are produced for simulated use testing and clinical evaluation. Another, a validation review, is conducted prior to transfer of the design to production. Generally, they are used to provide assurance that an activity or phase has been completed in an acceptable manner, and that the next activity or phase can begin.
As the figure illustrates, design validation encompasses verification and extends the assessment to address whether devices produced in accordance with the design actually satisfy user needs and intended uses.
An analogy to automobile design and development may help to clarify these concepts. Fuel efficiency is a common design requirement. This requirement might be expressed as the number of miles-per-gallon of a particular grade of gasoline for a specified set of driving conditions. As the design of the car proceeds, the requirements, including the one for fuel efficiency, are converted into the many layers of system and subsystem specifications needed for design. As these various systems and subsystems are designed, design verification methods are used to establish conformance of each design to its own specifications. Because several specifications directly affect fuel efficiency, many of the verification activities help to provide confirmation that the overall design will meet the fuel efficiency requirement. This might include simulated road testing of prototypes or actual road testing. This is establishing by objective evidence that the design output conforms to the fuel efficiency requirement. However, these verification activities alone are not sufficient to validate the design. The design may be validated when a representative sample of users have driven production vehicles under a specified range of driving conditions and judged the fuel efficiency to be adequate. This is providing objective evidence that the particular requirement for a specific intended use can be consistently fulfilled.
CONCURRENT ENGINEERING. Although the waterfall model is a useful tool for introducing design controls, its usefulness in practice is limited. The model does apply to the development of some simpler devices. However, for more complex devices, a concurrent engineering model is more representative of the design processes in use in the industry.
In a traditional waterfall development scenario, the engineering department completes the product design and formally transfers the design to production. Subsequently, other departments or organizations develop processes to manufacture and service the product. Historically, there has frequently been a divergence between the intent of the designer and the reality of the factory floor, resulting in such undesirable outcomes as low manufacturing yields, rework or redesign of the product, or unexpectedly high cost to service the product.
One benefit of concurrent engineering is the involvement of production and service personnel throughout the design process, assuring the mutual optimization of the characteristics of a device and its related processes. While the primary motivations of concurrent engineering are shorter development time and reduced production cost, the practical result is often improved product quality.
Concurrent engineering encompasses a range of practices and techniques. From a design control standpoint, it is sufficient to note that concurrent engineering may blur the line between development and production. On the one hand, the concurrent engineering model properly emphasizes that the development of production processes is a design rather than a manufacturing activity. On the other hand, various components of a design may enter production before the design as a whole has been approved. Thus, concurrent engineering and other more complex models of development usually require a comprehensive matrix of reviews and approvals to ensure that each component and process design is validated prior to entering production, and the product as a whole is validated prior to design release.
RISK MANAGEMENT AND DESIGN CONTROLS. Risk management is the systematic application of management policies, procedures, and practices to the tasks of identifying, analyzing, controlling, and monitoring risk. It is intended to be a framework within which experience, insight, and judgment are applied to successfully manage risk. It is included in this guidance because of its effect on the design process.
Risk management begins with the development of the design input requirements. As the design evolves, new risks may become evident. To systematically identify and, when necessary, reduce these risks, the risk management process is integrated into the design process. In this way, unacceptable risks can be identified and managed earlier in the design process when changes are easier to make and less costly.
An example of this is an exposure control system for a general purpose x-ray system. The control function was allocated to software. Late in the development process, risk analysis of the system uncovered several failure modes that could result in overexposure to the patient. Because the problem was not identified until the design was near completion, an expensive, independent, back-up timer had to be added to monitor exposure times.
THE QUALITY SYSTEM AND DESIGN CONTROLS. In addition to procedures and work instructions necessary for the implementation of design controls, policies and procedures may also be needed for other determinants of device quality that should be considered during the design process. The need for policies and procedures for these factors is dependent upon the types of devices manufactured by a company and the risks associated with their use. Management with executive responsibility has the responsibility for determining what is needed.
Example of topics for which policies and procedures may be appropriate are:
- risk management
- device reliability
- device durability
- device maintainability
- device serviceability
- human factors engineering
- software engineering
- use of standards
- configuration management
- compliance with regulatory requirements
- device evaluation (which may include third party product certification or approval)
- clinical evaluations
- document controls
- use of consultants
- use of subcontractors
- use of company historical data
SECTION A. GENERAL
I. REQUIREMENTS
§ 820.30(a) General.
- Each manufacturer of any class III or class II device, and the class I devices listed in paragraph (a) (2) of this section, shall establish and maintain procedures to control the design of the device in order to ensure that specified design requirements are met.
- The following class I devices are subject to design controls:
(i) Devices automated with computer software; and
(ii) The devices listed in the chart below.
Section | Device | 868.6810 | Catheter, Tracheobronchial Suction | 878.4460 | Glove, Surgeon's | 880.6760 | Restraint, Protective | 892.5650 | System, Applicator, Radionuclide, Manual | 892.5740 | Source, Radionuclide Teletherapy |
II. DEFINITIONS
§ 820.3 (n) Management with executive responsibility means those senior employees of a manufacturer who have the authority to establish or make changes to the manufacturer's quality policy and quality system.
§ 820.3 (s) Quality means the totality of features and characteristics that bear on the ability of a device to satisfy fitness-for-use, including safety and performance.
§ 820.3 (v) Quality system means the organizational structure, responsibilities, procedures, processes, and resources for implementing quality management.
Cross reference to ISO 9001:1994 and ISO/DIS 13485 Section 4.4.1 General.
III. DISCUSSION AND POINTS TO CONSIDER
The essential quality aspects and the regulatory requirements, such as safety, performance, and dependability of a product (whether hardware, software, services, or processed materials) are established during the design and development phase. Deficient design can be a major cause of quality problems.
The context within which product design is to be carried out should be set by the manufacturer's senior management. It is their responsibility to establish a design and development plan which sets the targets to be met. This plan defines the constraints within which the design is to be implemented.
The quality system requirements do not dictate the types of design process that a manufacturer must use. Manufacturers should use processes best suited to their needs. However, whatever the processes may be, it is important that the design controls are applied in an appropriate manner. This guidance document contains examples of how this might be achieved in a variety of situations.
It is important to note that the design function may apply to various facets of the operation having differing styles and time scales. Such facets are related to products, including services and software, as well as to their manufacturing processes.
Senior management needs to decide how the design function is to be managed and by whom. Senior management should also ensure that internal policies are established for design issues such as:
- assessing new product ideas
- training and retraining of design managers and design staff
- use of consultants
- evaluation of the design process
- product evaluation, including third party product certification and approvals
- patenting or other means of design protection
It is for senior management to ensure that adequate resources are available to carry out the design in the required time. This may involve reinforcing the skills and equipment available internally and/or obtaining external resources.
SECTION B. DESIGN AND DEVELOPMENT PLANNING
I. REQUIREMENTS
§ 820.30(b) Design and development planning.
- Each manufacturer shall establish and maintain plans that describe or reference the design and development activities and define responsibility for implementation.
- The plans shall identify and describe the interfaces with different groups or activities that provide, or result in, input to the design and development process.
- The plans shall be reviewed, updated, and approved as design and development evolves.
Cross-reference to ISO 9001:1994 and ISO/DIS 13485 sections 4.4.2 Design and development planning and 4.4.3 Organizational and technical interfaces.
II. DISCUSSION AND POINTS TO CONSIDER
Design and development planning is needed to ensure that the design process is appropriately controlled and that device quality objectives are met. The plans must be consistent with the remainder of the design control requirements. The following elements would typically be addressed in the design and development plan or plans:
- Description of the goals and objectives of the design and development program; i.e., what is to be developed;
- Delineation of organizational responsibilities with respect to assuring quality during the design and development phase, to include interface with any contractors;
- Identification of the major tasks to be undertaken, deliverables for each task, and individual or organizational responsibilities (staff and resources) for completing each task;
- Scheduling of major tasks to meet overall program time constraints;
- Identification of major reviews and decision points;
- Selection of reviewers, the composition of review teams, and procedures to be followed by reviewers;
- Controls for design documentation;
- Notification activities.
Planning enables management to exercise greater control over the design and development process by clearly communicating policies, procedures, and goals to members of the design and development team, and providing a basis for measuring conformance to quality system objectives.
Design activities should be specified at the level of detail necessary for carrying out the design process. The extent of design and development planning is dependent on the size of the developing organization and the size and complexity of the product to be developed. Some manufacturers may have documented policies and procedures which apply to all design and development activities. For each specific development program, such manufacturers may also prepare a plan which spells out the project-dependent elements in detail, and incorporates the general policies and procedures by reference. Other manufacturers may develop a comprehensive design and development plan which is specifically tailored to each individual project.
In summary, the form and organization of the planning documents are less important than their content. The following paragraphs discuss the key elements of design and development planning.
ORGANIZATIONAL RESPONSIBILITIES. The management responsibility section of the quality system requirements requires management to establish a quality policy and implement an organizational structure to ensure quality. These are typically documented in a quality manual or similarly named document. In some cases, however, the design and development plan, rather than the quality manual, is the best vehicle for describing organizational responsibilities relative to design and development activities. The importance of defining responsibilities with clarity and without ambiguity should be recognized. When input to the design is from a variety of sources, their interrelationships and interfaces (as well as the pertinent responsibilities and authorities) should be defined, documented, coordinated, and controlled. This might be the case, for example, if a multidisciplinary product development team is assembled for a specific project, or if the team includes suppliers, contract manufacturers, users, outside consultants, or independent auditors.
TASK BREAKDOWN. The plan establishes, to the extent possible:
- The major tasks required to develop the product
- The time involved for each major task
- The resources and personnel required
- The allocation of responsibilities for completing each major task
- The prerequisite information necessary to start each major task and the interrelationship between tasks
- The form of each task output or deliverable
- Constraints, such as applicable codes, standards, and regulations
Tasks for all significant design activities, including verification and validation tasks, should be included in the design and development plan. For example, if clinical trials are anticipated, there may be tasks associated with appropriate regulatory requirements.
For complex projects, rough estimates may be provided initially, with the details left for the responsible organizations to develop. As development proceeds, the plan should evolve to incorporate more and better information.
The relationships between tasks should be presented in such a way that they are easily understood. It should be clear which tasks depend on others, and which tasks need to be performed concurrently. Planning should reflect the degree of perceived development risk; for example, tasks involving new technology or processes should be spelled out in greater detail, and perhaps be subjected to more reviews and checks, than tasks which are perceived as routine or straightforward.
The design and development plan may include a schedule showing starting and completion dates for each major task, project milestone, or key decision points. The method chosen and the detail will vary depending on the complexity of the project and the level of risk associated with the device. For small projects, the plan may consist of only a simple flow diagram or computer spreadsheet. For larger projects, there are a number of project management tools that are used to develop plans. Three of the most commonly used are the Program Evaluation and Review Technique (PERT), the Critical Path Method (CPM), and the Gantt chart. Software is available in many forms for these methods. When selecting these tools, be careful to choose one that best fits the needs of the project. Some of the software programs are far more complex than may be necessary.
Unless a manufacturer has experience with the same type of device, the plan will initially be limited in scope and detail. As work proceeds, the plan is refined. Lack of experience in planning often leads to optimistic schedules, but slippage may also occur for reasons beyond the control of planners, for example, personnel turnover, materiel shortage, or unexpected problems with a design element or process. Sometimes the schedule can be compressed by using additional resources, such as diverting staff or equipment from another project, hiring a contractor, or leasing equipment.
It is important that the schedule be updated to reflect current knowledge. At all times, the plan should be specified at a level of detail enabling management to make informed decisions, and provide confidence in meeting overall schedule and performance objectives. This is important because scheduling pressures have historically been a contributing factor in many design defects which caused injury. To the extent that good planning can prevent schedule pressures, the potential for design errors is reduced.
However, no amount of planning can eliminate all development risk. There is inherent conflict between the desire to maximize performance and the need to meet business objectives, including development deadlines. In some corporate cultures, impending deadlines create enormous pressure to cut corners. Planning helps to combat this dilemma by ensuring management awareness of pressure points. With awareness, decisions are more likely to be made with appropriate oversight and consideration of all relevant factors. Thus, when concessions to the clock must be made, they can be justified and supported.
SECTION C. DESIGN INPUT
I. REQUIREMENTS
§ 820.30(c) Design input.
- Each manufacturer shall establish and maintain procedures to ensure that the design requirements relating to a device are appropriate and address the intended use of the device, including the needs of the user and patient.
- The procedures shall include a mechanism for addressing incomplete, ambiguous, or conflicting requirements.
- The design input requirements shall be documented and shall be reviewed and approved by designated individual(s).
- The approval, including the date and signature of the individual(s) approving the requirements, shall be documented.
Cross reference to ISO 9001:1994 and ISO/DIS 13485 section 4.4.4 Design input.
II. DEFINITIONS
§ 820.3(f) Design input means the physical and performance requirements of a device that are used as a basis for device design.
III. DISCUSSION AND POINTS TO CONSIDER
Design input is the starting point for product design. The requirements which form the design input establish a basis for performing subsequent design tasks and validating the design. Therefore, development of a solid foundation of requirements is the single most important design control activity.
Many medical device manufacturers have experience with the adverse effects that incomplete requirements can have on the design process. A frequent complaint of developers is that "there's never time to do it right, but there's always time to do it over." If essential requirements are not identified until validation, expensive redesign and rework may be necessary before a design can be released to production.
By comparison, the experience of companies that have designed devices using clear-cut, comprehensive sets of requirements is that rework and redesign are significantly reduced and product quality is improved. They know that the development of requirements for a medical device of even moderate complexity is a formidable, time-consuming task. They accept the investment in time and resources required to develop the requirements because they know the advantages to be gained in the long run.
Unfortunately, there are a number of common misconceptions regarding the meaning and practical application of the quality system requirements for design input. Many seem to arise from interpreting the requirements as a literal prescription, rather than a set of principles to be followed. In this guidance document, the focus is on explaining the principles and providing examples of how they may be applied in typical situations.
CONCEPT DOCUMENTS VERSUS DESIGN INPUT In some cases, the marketing staff, who maintain close contact with customers and users, determine a need for a new product, or enhancements to an existing product. Alternatively, the idea for a new product may evolve out of a research or clinical activity. In any case, the result is a concept document specifying some of the desired characteristics of the new product.
Some members of the medical device community view these marketing memoranda, or the equivalent, as the design input. However, that is not the intent of the quality system requirements. Such concept documents are rarely comprehensive, and should not be expected to be so. Rather, the intent of the quality system requirements is that the product conceptual description be elaborated, expanded, and transformed into a complete set of design input requirements which are written to an engineering level of detail.
This is an important concept. The use of qualitative terms in a concept document is both appropriate and practical. This is often not the case for a document to be used as a basis for design. Even the simplest of terms can have enormous design implications. For example, the term "must be portable" in a concept document raises questions in the minds of product developers about issues such as size and weight limitations, resistance to shock and vibration, the need for protection from moisture and corrosion, the capability of operating over a wide temperature range, and many others. Thus, a concept document may be the starting point for development, but it is not the design input requirement. This is a key principle-the design input requirements are the result of the first stage of the design control process.
RESEARCH AND DEVELOPMENT. Some manufacturers have difficulty in determining where research ends and development begins. Research activities may be undertaken in an effort to determine new business opportunities or basic characteristics for a new product. It may be reasonable to develop a rapid prototype to explore the feasibility of an idea or design approach, for example, prior to developing design input requirements. But manufacturers should avoid falling into the trap of equating the prototype design with a finished product design. Prototypes at this stage lack safety features and ancillary functions necessary for a finished product, and are developed under conditions which preclude adequate consideration of product variability due to manufacturing.
RESPONSIBILITY FOR DESIGN INPUT DEVELOPMENT. Regardless of who developed the initial product concept, product developers play a key role in developing the design input requirements. When presented with a set of important characteristics, it is the product developers who understand the auxiliary issues that must be addressed, as well as the level of detail necessary to design a product. Therefore, a second key principle is that the product developer(s) ultimately bear responsibility for translating user and/or patient needs into a set of requirements which can be validated prior to implementation. While this is primarily an engineering function, the support or full participation of production and service personnel, key suppliers, etc., may be required to assure that the design input requirements are complete.
Care must be exercised in applying this principle. Effective development of design input requirements encompasses input from both the product developer as well as those representing the needs of the user, such as marketing. Terminology can be a problem. In some cases, the product conceptual description may be expressed in medical terms. Medical terminology is appropriate in requirements when the developers and reviewers are familiar with the language, but it is often preferable to translate the concepts into engineering terms at the requirements stage to minimize miscommunication with the development staff.
Another problem is incorrect assumptions. Product developers make incorrect assumptions about user needs, and marketing personnel make incorrect assumptions about the needs of the product designers. Incorrect assumptions can have serious consequences that may not be detected until late in the development process. Therefore, both product developers and those representing the user must take responsibility for critically examining proposed requirements, exploring stated and implied assumptions, and uncovering problems.
Some examples should clarify this point. A basic principle is that design input requirements should specify what the design is intended to do while carefully avoiding specific design solutions at this stage. For example, a concept document might dictate that the product be housed in a machined aluminum case. It would be prudent for product developers to explore why this type of housing was specified. Perhaps there is a valid reason-superior electrical shielding, mechanical strength, or reduced time to market as compared to a cast housing. Or perhaps machined aluminum was specified because a competitor's product is made that way, or simply because the user didn't think plastic would be strong enough.
Not all incorrect assumptions are made by users. Incorrect assumptions made by product developers may be equally damaging. Failure to understand the abuse to which a portable instrument would be subjected might result in the selection of housing materials inadequate for the intended use of the product.
There are occasions when it may be appropriate to specify part of the design solution in the design input requirements. For example, a manufacturer may want to share components or manufacturing processes across a family of products in order to realize economies of scale, or simply to help establish a corporate identity. In the case of a product upgrade, there may be clear consensus regarding the features to be retained. However, it is important to realize that every such design constraint reduces implementation flexibility and should therefore be documented and identified as a possible conflicting requirement for subsequent resolution.
SCOPE AND LEVEL OF DETAIL. Design input requirements must be comprehensive. This may be quite difficult for manufacturers who are implementing a system of design controls for the first time. Fortunately, the process gets easier with practice. It may be helpful to realize that design input requirements fall into three categories. Virtually every product will have requirements of all three types.
- Functional requirements specify what the device does, focusing on the operational capabilities of the device and processing of inputs and the resultant outputs.
- Performance requirements specify how much or how well the device must perform, addressing issues such as speed, strength, response times, accuracy, limits of operation, etc. This includes a quantitative characterization of the use environment, including, for example, temperature, humidity, shock, vibration, and electromagnetic compatibility. Requirements concerning device reliability and safety also fit into this category.
- Interface requirements specify characteristics of the device which are critical to compatibility with external systems; specifically, those characteristics which are mandated by external systems and outside the control of the developers. One interface which is important in every case is the user and/or patient interface.
What is the scope of the design input requirements development process and how much detail must be provided? The scope is dependent upon the complexity of a device and the risk associated with its use. For most medical devices, numerous requirements encompassing functions, performance, safety, and regulatory concerns are implied by the application. These implied requirements should be explicitly stated, in engineering terms, in the design input requirements.
Determining the appropriate level of detail requires experience. However, some general guidance is possible. The marketing literature contains product specifications, but these are superficial. The operator and service manuals may contain more detailed specifications and performance limits, but these also fall short of being comprehensive. Some insight as to what is necessary is provided by examining the requirements for a very common external interface. For the power requirements for AC-powered equipment, it is not sufficient to simply say that a unit shall be AC-powered. It is better to say that the unit shall be operable from AC power in North America, Europe, and Japan, but that is still insufficient detail to implement or validate the design. If one considers the situation just in North America, where the line voltage is typically 120 volts, many systems are specified to operate over the range of 108 to 132 volts. However, to account for the possibility of brownout, critical devices may be specified to operate from 95 to 132 volts or even wider ranges. Based on the intended use of the device, the manufacturer must choose appropriate performance limits.
There are many cases when it is impractical to establish every functional and performance characteristic at the design input stage. But in most cases, the form of the requirement can be determined, and the requirement can be stated with a to-be-determined (TBD) numerical value or a range of possible values. This makes it possible for reviewers to assess whether the requirements completely characterize the intended use of the device, judge the impact of omissions, and track incomplete requirements to ensure resolution.
For complex designs, it is not uncommon for the design input stage to consume as much as thirty percent of the total project time. Unfortunately, some managers and developers have been trained to measure design progress in terms of hardware built, or lines of software code written. They fail to realize that building a solid foundation saves time during the implementation. Part of the solution is to structure the requirements documents and reviews such that tangible measures of progress are provided.
At the other extreme, many medical devices have very simple requirements. For example, many new devices are simply replacement parts for a product, or are kits of commodity items. Typically, only the packaging and labeling distinguishes these products from existing products. In such cases, there is no need to recreate the detailed design input requirements of the item. It is acceptable to simply cite the predecessor product documentation, add any new product information, and establish the unique packaging and labeling requirements.
ASSESSING DESIGN INPUT REQUIREMENTS FOR ADEQUACY. Eventually, the design input must be reviewed for adequacy. After review and approval, the design input becomes a controlled document. All future changes will be subject to the change control procedures, as discussed in Section I (Design Changes).
Any assessment of design input requirements boils down to a matter of judgment. As discussed in Section E (Design Review), it is important for the review team to be multidisciplinary and to have the appropriate authority. A number of criteria may be employed by the review team.
- Design input requirements should be unambiguous. That is, each requirement should be able to be verified by an objective method of analysis, inspection, or testing. For example, it is insufficient to state that a catheter must be able to withstand repeated flexing. A better requirement would state that the catheter should be formed into a 50 mm diameter coil and straightened out for a total of fifty times with no evidence of cracking or deformity. A qualified reviewer could then make a judgment whether this specified test method is representative of the conditions of use.
- Quantitative limits should be expressed with a measurement tolerance. For example, a diameter of 3.5 mm is an incomplete specification. If the diameter is specified as 3.500±0.005 mm, designers have a basis for determining how accurate the manufacturing processes have to be to produce compliant parts, and reviewers have a basis for determining whether the parts will be suitable for the intended use.
- The set of design input requirements for a product should be self-consistent. It is not unusual for requirements to conflict with one another or with a referenced industry standard due to a simple oversight. Such conflicts should be resolved early in the development process.
- The environment in which the product is intended to be used should be properly characterized. For example, manufacturers frequently make the mistake of specifying "laboratory" conditions for devices which are intended for use in the home. Yet, even within a single country, relative humidity in a home may range from 20 percent to 100 percent (condensing) due to climactic and seasonal variations. Household temperatures in many climates routinely exceed 40 °C during the hot season. Altitudes may exceed 3,000 m, and the resultant low atmospheric pressure may adversely affect some kinds of medical equipment. If environmental conditions are fully specified, a qualified reviewer can make a determination of whether the specified conditions are representative of the intended use.
- When industry standards are cited, the citations should be reviewed for completeness and relevance. For example, one medical device manufacturer claimed compliance with an industry standard covering mechanical shock and vibration. However, when the referenced standard was examined by a reviewer, it was found to prescribe only the method of testing, omitting any mention of pass/fail criteria. It was incumbent on the manufacturer in this case to specify appropriate performance limits for the device being tested, as well as the test method.
EVOLUTION OF THE DESIGN INPUT REQUIREMENTS. Large development projects often are implemented in stages. When this occurs, the design input requirements at each stage should be developed and reviewed following the principles set forth in this section. Fortunately, the initial set of requirements, covering the overall product, is by far the most difficult to develop. As the design proceeds, the output from the early stages forms the basis for the subsequent stages, and the information available to designers is inherently more extensive and detailed.
It is almost inevitable that verification activities will uncover discrepancies which result in changes to the design input requirements. There are two points to be made about this. One is that the change control process for design input requirements must be carefully managed. Often, a design change to correct one problem may create a new problem which must be addressed. Throughout the development process, it is important that any changes are documented and communicated to developers so that the total impact of the change can be determined. The change control process is crucial to device quality.
The second point is that extensive rework of the design input requirements suggests that the design input requirements may not be elaborated to a suitable level of detail, or insufficient resources are being devoted to defining and reviewing the requirements. Managers can use this insight to improve the design control process. From a design control perspective, the number of requirements changes made is less important than the thoroughness of the change control process.
SECTION D. DESIGN OUTPUT
I. REQUIREMENTS
§ 820.30(d) Design output.
- Each manufacturer shall establish and maintain procedures for defining and documenting design output in terms that allow an adequate evaluation of conformance to design input requirements.
- Design output procedures shall contain or make reference to acceptance criteria and shall ensure that those design outputs that are essential for the proper functioning of the device are identified.
- Design output shall be documented, reviewed, and approved before release.
- The approval, including the date and signature of the individual(s) approving the output, shall be documented.
Cross-reference to ISO 9001:1994 and ISO/DIS 13485 section 4.4.5 Design output.
II. DEFINITIONS
§ 820.3(g) Design output means the results of a design effort at each design phase and at the end of the total design effort. The finished design output is the basis for the device master record. The total finished design output consists of the device, its packaging and labeling, and the device master record.
§ 820.3(y) Specification means any requirement with which a product, process, service, or other activity must conform.
III. DISCUSSION AND POINTS TO CONSIDER
The quality system requirements for design output can be separated into two elements: Design output should be expressed in terms that allow adequate assessment of conformance to design input requirements and should identify the characteristics of the design that are crucial to the safety and proper functioning of the device. This raises two fundamental issues for developers:
- What constitutes design output?
- Are the form and content of the design output suitable?
The first issue is important because the typical development project produces voluminous records, some of which may not be categorized as design output. On the other hand, design output must be reasonably comprehensive to be effective. As a general rule, an item is design output if it is a work product, or deliverable item, of a design task listed in the design and development plan, and the item defines, describes, or elaborates an element of the design implementation. Examples include block diagrams, flow charts, software high-level code, and system or subsystem design specifications. The design output in one stage is often part of the design input in subsequent stages.
Design output includes production specifications as well as descriptive materials which define and characterize the design.
PRODUCTION SPECIFICATIONS. Production specifications include drawings and documents used to procure components, fabricate, test, inspect, install, maintain, and service the device, such as the following:
- assembly drawings
- component and material specifications
- production and process specifications
- software machine code (e.g., diskette or master EPROM)
- work instructions
- quality assurance specifications and procedures
- installation and servicing procedures
- packaging and labeling specifications, including methods and processes used
In addition, as discussed in Section H (Design Transfer), production specifications may take on other forms. For example, some manufacturers produce assembly instructions on videotapes rather than written instructions. Similarly, a program diskette, used by a computer-aided milling machine to fabricate a part, would be considered a production specification. The videotape and the software on the program diskette are part of the device master record.
OTHER DESCRIPTIVE MATERIALS. Other design output items might be produced which are necessary to establish conformance to design input requirements, but are not used in its production. For example, for each part which is fabricated by computer-aided machine, there should be an assembly drawing which specifies the dimensions and characteristics of the part. It is a part of the design output because it establishes the basis for the machine tool program used to fabricate the part. Other examples of design output include the following:
- the results of risk analysis
- software source code
- results of verification activities
- biocompatibility test results
- bioburden test results
FORM AND CONTENT. Manufacturers must take steps to assure that the design output characterizes all important aspects of the design and is expressed in terms which allow adequate verification and validation. Two basic mechanisms are available to manufacturers to accomplish these objectives.
- First, the manufacturer proactively can specify the form and content of design output at the planning stage. For some types of design output, form and content may be codified in a consensus standard which can be referenced. In other cases, a manufacturer could specify the desired characteristics, or even simply specify that the form and content of an existing document be followed.
- Second, form and content can be reviewed retroactively as a part of the design verification process. For example, the verification of design output could include assessing whether specified documentation standards have been adhered to.
As these examples illustrate, conformance with the quality system requirements concerning design output generally requires no "extra" effort on the part of the manufacturer, but simply the application of some common sense procedures during the planning, execution, and review of design tasks.
SECTION E. DESIGN REVIEW
I. REQUIREMENTS
§ 820.30(e) Design review.
- Each manufacturer shall establish and maintain procedures to ensure that formal documented reviews of the design results are planned and conducted at appropriate stages of the device's design development.
- The procedures shall ensure that participants at each design review include representatives of all functions concerned with the design stage being reviewed and an individual(s) who does not have direct responsibility for the design stage being reviewed, as well as any specialists needed.
- The results of a design review, including identification of the design, the date, and the individual(s) performing the review, shall be documented in the design history file (the DHF).
Cross-reference to ISO 9001:1994 and ISO/DIS 13485 section 4.4.6 Design review.
II. DEFINITIONS
§ 820.3(h) Design review means a documented, comprehensive, systematic examination of a design to evaluate the adequacy of the design requirements, to evaluate the capability of the design to meet these requirements, and to identify problems.
III. DISCUSSION AND POINTS TO CONSIDER
In general, formal design reviews are intended to:
- provide a systematic assessment of design results, including the device design and the associated designs for production and support processes;
- provide feedback to designers on existing or emerging problems;
- assess project progress; and/or
- provide confirmation that the project is ready to move on to the next stage of development.
Many types of reviews occur during the course of developing a product. Reviews may have both an internal and external focus. The internal focus is on the feasibility of the design and the produceability of the design with respect to manufacturing and support capabilities. The external focus is on the user requirements; that is, the device design is viewed from the perspective of the user.
The nature of reviews changes as the design progresses. During the initial stages, issues related to design input requirements will predominate. Next, the main function of the reviews may be to evaluate or confirm the choice of solutions being offered by the design team. Then, issues such as the choice of materials and the methods of manufacture become more important. During the final stages, issues related to the verification, validation, and production may predominate.
The term "review" is commonly used by manufacturers to describe a variety of design assessment activities. Most, but not all, of these activities meet the definition of formal design reviews. The following exceptions may help to clarify the distinguishing characteristics of design reviews.
- Each design document which constitutes the formal output, or deliverable, of a design task is normally subject to evaluation activities, sometimes referred to as informal peer review, supervisory review, or technical assessment. These activities, while they may be called reviews, are often better described as verification activities, because they are not intended to be comprehensive, definitive, and multidisciplinary in their scope. Rather, their purpose is to confirm that design output meets design input. Verification activities affect and add to the design output, and are themselves subject to subsequent design review.
- Developers may conduct routine or ad hoc meetings to discuss an issue, coordinate activities, or assess development progress. Decisions from such meetings may not require formal documentation; however, if a significant issue is resolved, this should be documented. If the outcome results in change to an approved design document, then applicable change control procedures should be followed, as discussed in Section I (Design Changes).
Control of the design review process is achieved by developing and implementing a formal design review program consistent with quality system requirements. The following issues should be addressed and documented in the design and development plan(s).
NUMBER AND TYPE OF REVIEWS. It is a well-accepted fact that the cost to correct design errors increases as the design nears completion, and the flexibility to implement an optimal solution decreases. When an error is discove
|
|