[现场管理] 关于灭菌柜真空泄漏测试方法的疑惑

[复制链接]
药徒
发表于 2019-5-1 18:04:15 | 显示全部楼层 |阅读模式

欢迎您注册蒲公英

您需要 登录 才可以下载或查看,没有帐号?立即注册

x
公司有新华灭菌柜的人应该都知道,他们家的真空泄漏测试方法是抽真空到-90kpa,维持15min,压降不超过2kpa,即视为合格。
按照EN285的要求,这个方法是不符合要求的。

按EN285的要求,要抽真空至绝对压力7kpa也就是-94.3kpa,维持10min,标准为≤0.13kpa/min。
18.3.3 With the temperature stabilized and the sterilizer chamber empty except for fixed furniture and
necessary sensors, start the test cycle. When the absolute pressure in the sterilizer chamber is 7 kPa or
below close all the valves connected to the sterilizer chamber and stop the vacuum pump. Observe and
record the time (t1) and the pressure (p1). Wait at least 300 s and not more than 600 s to allow
evaporation of condensate in the sterilizer chamber and then observe and record the pressure (p2) in
the sterilizer chamber and the time (t2). After a further (600 ± 10) s, again observe and record the
pressure (p3) and the time (t3).

此外EN285对仪器仪表的要求是  压力表精度应为0.1kpa,秒表应为15min内误差不超过0.5秒。  
新华的PLC时间是1h内不超过36秒。

如果按EN285改真空值,泵的性能也很难达到要求。
所以问题是:是否必须抽真空至-94.3kpa,PLC的时间精度是否需要符合EN 285的要求。

回复

使用道具 举报

药徒
发表于 2019-5-5 09:43:35 | 显示全部楼层
就一个问题,新华灭菌柜按照哪些标准来制造的,符合中国GMP要求就行了
回复

使用道具 举报

药徒
发表于 2020-5-17 19:53:40 | 显示全部楼层
换一个好一点的泵,-95kpa都不是问题
回复

使用道具 举报

药神
发表于 2022-7-10 20:37:00 | 显示全部楼层
非常感谢分享
回复

使用道具 举报

您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

×发帖声明
1、本站为技术交流论坛,发帖的内容具有互动属性。您在本站发布的内容:
①在无人回复的情况下,可以通过自助删帖功能随时删除(自助删帖功能关闭期间,可以联系管理员微信:8542508 处理。)
②在有人回复和讨论的情况下,主题帖和回复内容已构成一个不可分割的整体,您将不能直接删除该帖。
2、禁止发布任何涉政、涉黄赌毒及其他违反国家相关法律、法规、及本站版规的内容,详情请参阅《蒲公英论坛总版规》。
3、您在本站发表、转载的任何作品仅代表您个人观点,不代表本站观点。不要盗用有版权要求的作品,转贴请注明来源,否则文责自负。
4、请认真阅读上述条款,您发帖即代表接受上述条款。

QQ|手机版|蒲公英|ouryao|蒲公英 ( 京ICP备14042168号-1 )  京ICP证150354号  互联网药品信息服务证书编号: (京)-非经营性-2024-0033

GMT+8, 2024-11-10 18:33

Powered by Discuz! X3.4运维单位:苏州豚鼠科技有限公司

Copyright © 2001-2020, Tencent Cloud.

声明:蒲公英网站所涉及的原创文章、文字内容、视频图片及首发资料,版权归作者及蒲公英网站所有,转载要在显著位置标明来源“蒲公英”;禁止任何形式的商业用途。违反上述声明的,本站及作者将追究法律责任。
快速回复 返回顶部 返回列表