8.2.1.2.3 Throughout the holding time the temperature measured at the reference measurement point of the sterilizer chamber, any temperature measured within the test pack and the theoretical temperature of the saturated steam determined from the measured sterilizer chamber pressure shall:
a) be within the sterilization temperature band;
b) not differ from another by more than 2 K.