KF has a high accuracy and precision, typically within 1% of available water, e.g. 3.00% appears as 2.97 - 3.03%. Although KF is a destructive analysis, the sample quantity is small and is typically limited by the accuracy of weighing. For example, in order to obtain an accuracy of 1% using a scale with the typical accuracy of 0.2 mg, the sample must contain 20 mg water, which is e.g. 200 mg for a sample with 10% water. For coulometers, the measuring range is from 1-5 ppm to ca. 5%. Volumetric KF readily measures samples up to 100%, but requires impractically large amounts of sample for analytes with less than 0.05% water.仅供参考!
https://en.wikipedia.org/wiki/Karl_Fischer_titration
|
|