EN 285:2006+A2:2009, 8.2.1.2.1(=8.2.1.3.1) The equilibration time shall not exceed 15 s for sterilizer chambers up to 800 l usable space and 30 s for larger sterilizer chambers.
ISO 17665.2-2009, A.4.1.2 b) The equilibration time should not exceed 15 s for sterilizer chambers up to 800 l usable space and 30 s for larger sterilizer chambers.
GB 8599-2008与EN 285和ISO 17665.2-2009相同。
EN 13060:2004+A1:2009, 4.6.4 Equilibration time
The equilibration time shall not exceed 15 s.
An equilibration time not exceeding 30 s is acceptable if:
- the rise of the theoretical steam temperature during the last 10 K of the heating stage is less than 8 K/min but greater than 1 K/min;
- during the last 10 K of the heating stage all temperatures measured in the chamber and the load as well as the theoretical steam temperature do not differ from one another by more than 2 K.
|
|